Symptom: The observed retention times of the peptides eluting are compressed, are not as spread out as expected, suggesting the elution gradient is compressed
Test: Use PepCalMix LC-MS test to check the peptide elution time and spread
Solution:
- Make sure you are performing the recommended regular maintenance tests on your low flow LC system
- Ensure there are no leaks in your flow path
- For the SCIEX low flow LC systems:
- Re-initialize pressure transducers regularly (at a minimum, biweekly is recommended for best performance)
- Perform a leak test on your LC pump. If this fails, contact your service representative
- If the leak test passes, check flow module calibration on your LC pump, recalibrate if necessary. If this does not help or the calibration is off by more than 20%, contact your service representative
Symptom: The elution of peptides is delayed, appearing later than expected.
Test: Use PepCalMix LC-MS test to check the peptide elution time and spread, typical chromatograms are shown in the protocol.
Solution:
- Check for leaks in your gradient flow path, both pre- and post-column. Check the pressure in the gradient pump to see if it is lower than expected.
- Dead volume in pre-column flow path will create a gradient delay, check tubing lengths and connections for excess dead volume