Standard front-end cleaning for TripleTOF systems by Christie Hunter | 0 CommentsGeneral points: Use of laboratory deionized (DI) water is specified at points in the procedure. However, if there are any concerns regarding quality, please utilize a reputable supplier of pure deionized water (>18.2M ohm). 5-10 liters will typically be required...
Simple tests for determining whether to clean your QJet ion guide and IQ0 Lens on your TripleTOF system by Christie Hunter | 0 CommentsAssessing the linear response of ITC (dynamic ion transmission control) is an easy way to determine whether you should clean your QJet ion guide and IQ0 lens. A clean rod set and lens should give a linear response when ramping the ITC parameter and the ratio of 132...
Standard front-end cleaning for TripleTOF systems by Christie Hunter | Instrument maintenance, QTOF, SCIEX information, TechnologyGeneral points: Use of laboratory deionized (DI) water is specified at points in the procedure. However, if there are any concerns regarding quality, please utilize a reputable supplier of pure deionized water (>18.2M ohm). 5-10 liters will typically be required...
Simple tests for determining whether to clean your QJet ion guide and IQ0 Lens on your TripleTOF system by Christie Hunter | Instrument maintenance, QTOF, SCIEX information, TechnologyAssessing the linear response of ITC (dynamic ion transmission control) is an easy way to determine whether you should clean your QJet ion guide and IQ0 lens. A clean rod set and lens should give a linear response when ramping the ITC parameter and the ratio of 132...