GEN-MKT-18-7897-A
Jun 3, 2025 | Blogs, Echo® MS+ system, Pharma | 0 comments
Read Time: 2 min
Depending on the samples that you are running on the system, it is possible for the Echo MS electrode to become dirty or occluded over time. Below are two different cleaning strategies that will be helpful for you to maintain your system and keep your electrodes running well.
This protocol is easy to perform between running your sample plates.
This protocol is an offline protocol that can be used periodically for a more intensive wash of the electrode or to recover electrodes that are suffering from reduced flow rates.
Connect the OPI electrode to a small union (1/16 to 1/32) , IDEX Part # P-881 )
Some electrodes may not be able to be recovered but we have had good success with this method.
As regulatory scrutiny increases and detection requirements tighten, laboratories are facing a new question: How can TFA be measured reliably, sensitively, and at scale?
Glycosylation is one of the more structurally diverse and biologically impactful PTMs in protein therapeutics. Both N‑linked and O‑linked glycans influence protein folding, stability, and biological activity. Given these effects on biotherapeutics, glycosylation is a closely monitored critical quality attribute (CQA). Comprehensive and site‑specific characterization of glycosylation is essential for informed decision‑making throughout drug discovery and development.