Methods for electrode cleaning for Echo MS systems

Jun 3, 2025 | Blogs, Echo® MS+ system, Pharma | 0 comments

Read Time: 2 min

Depending on the samples that you are running on the system, it is possible for the Echo MS electrode to become dirty or occluded over time.  Below are two different cleaning strategies that will be helpful for you to maintain your system and keep your electrodes running well.

  1. Quick cleaning protocol

This protocol is easy to perform between running your sample plates.

  • From the Maintenance menu of the Echo MS direct control, select OPI Port Access then Lift to provide access to the top of the OPI port.
  • Disconnect electrode by loosening the finger tight red nut and pulling straight up
  • Place the OPI port end of the electrode into solvent. With the MS system still active, the draw created from the GS1 flow will draw solvent through the OPI into the MS source, turn Spray voltage to 0.
  • Wash with Milli-Q water for 1 min
  • Wash with 20% Acetonitrile with 0.1 M Formic Acid (0.4% FA) for 20 min
    • Drawing speed is about 450 µL/min based on the consumption of 20% Acetonitrile with 0.1 M FA
  • Wash with Milli-Q water for 1 min
  • Reattach the electrode to the system and tighten the red nut finger tight.
  • Restart the mobile phase flow and check the Maximum Supported Flow (MSF)

 

  1. Intensive cleaning protocol

This protocol is an offline protocol that can be used periodically for a more intensive wash of the electrode or to recover electrodes that are suffering from reduced flow rates.

Connect the OPI electrode to a small union (1/16 to 1/32) , IDEX Part # P-881 )

  • Place the electrode end of the capillary in a waste vessel capable of holding 200 mL or more
  • Flow at least 200 ml of 50% methanol through the electrode at a flow rate of 5 – 8 mL/min using an HPLC pump
  • Test on Echo MS system for droplet linearity, reproducibility, and peak shape
  • If storing the recovered electrode flush with 100% methanol before storing.

Some electrodes may not be able to be recovered but we have had good success with this method.

The people behind the service: Building trust, solving problems, and keeping science moving

When customers think about service for their mass spectrometer or capillary electrophoresis (CE) instrument, they often think about repairs. But for SCIEX Field Service Engineer Tom Abbott, the role is much more than fixing instruments. It’s about troubleshooting complex problems, building trusted relationships, and helping labs stay productive.

The future of food testing is trust

Consumers expect more than safe food. They want confidence that the food they buy is authentic, contains what the label says it does, and meets their expectations for quality and origin.

How to evaluate LC-MS and CE system service providers: A practical guide

Selecting a maintenance, service, and support provider for liquid chromatography-mass spectrometry (LC-MS) and capillary electrophoresis (CE) systems can directly affect instrument uptime, analytical performance, compliance, and total cost of ownership. As service models evolve, your laboratory needs clear, defensible criteria to evaluate providers that support these critical systems throughout their lifecycle. 

Posted by

Trained and working in Organic Synthesis I started using mass spectrometry to aid in my synthesis work. As my interest in mass spectrometry grew, I change roles several times within the pharmaceutical industry to be able to focus on its use and applications. I joined SCIEX in 2006, initially as a demo chemist, then with R&D, primarily focused on Analyst, and eventually SCIEX OS software development. More recently I switched to technical product management for Echo MS systems.

0 Comments

Submit a Comment

Pin It on Pinterest

Share This

Share this post with your network