Methods for electrode cleaning for Echo MS systems

Jun 3, 2025 | Blogs, Echo® MS+ system, Pharma | 0 comments

Read Time: 2 min

Depending on the samples that you are running on the system, it is possible for the Echo MS electrode to become dirty or occluded over time.  Below are two different cleaning strategies that will be helpful for you to maintain your system and keep your electrodes running well.

  1. Quick cleaning protocol

This protocol is easy to perform between running your sample plates.

  • From the Maintenance menu of the Echo MS direct control, select OPI Port Access then Lift to provide access to the top of the OPI port.
  • Disconnect electrode by loosening the finger tight red nut and pulling straight up
  • Place the OPI port end of the electrode into solvent. With the MS system still active, the draw created from the GS1 flow will draw solvent through the OPI into the MS source, turn Spray voltage to 0.
  • Wash with Milli-Q water for 1 min
  • Wash with 20% Acetonitrile with 0.1 M Formic Acid (0.4% FA) for 20 min
    • Drawing speed is about 450 µL/min based on the consumption of 20% Acetonitrile with 0.1 M FA
  • Wash with Milli-Q water for 1 min
  • Reattach the electrode to the system and tighten the red nut finger tight.
  • Restart the mobile phase flow and check the Maximum Supported Flow (MSF)

 

  1. Intensive cleaning protocol

This protocol is an offline protocol that can be used periodically for a more intensive wash of the electrode or to recover electrodes that are suffering from reduced flow rates.

Connect the OPI electrode to a small union (1/16 to 1/32) , IDEX Part # P-881 )

  • Place the electrode end of the capillary in a waste vessel capable of holding 200 mL or more
  • Flow at least 200 ml of 50% methanol through the electrode at a flow rate of 5 – 8 mL/min using an HPLC pump
  • Test on Echo MS system for droplet linearity, reproducibility, and peak shape
  • If storing the recovered electrode flush with 100% methanol before storing.

Some electrodes may not be able to be recovered but we have had good success with this method.

Sterile but toxic: Cereulide detection in food matrices

In a recent webinar, Ming Gao, Ph.D., Senior Research Chemist at Mérieux NutriSciences, explored the growing challenge of cereulide contamination in food products and the analytical strategies needed to detect this heat-stable toxin at trace levels. The session examined how LC-MS/MS can help laboratories overcome the limitations of traditional microbiological screening and meet emerging regulatory requirements.

How SCIEX Now empowers mass spec users to work with more control, clarity and confidence

In this interview, Andy Wilcox, Senior Market Development Manager, Aftermarket and Software, speaks with Susanna Baqué, Senior Director of Global Customer Experience, about how SCIEX Now gives customers one connected digital workspace to manage instruments, reduce time spent searching for information, resolve issues faster and keep their laboratory running more efficiently and with greater confidence.

Bioanalysis across modalities: Advancing confident decisions with modern LC‑MS strategies

As therapeutic pipelines continue to diversify, bioanalysis is being asked to do more than ever before. From small molecules to complex biologics, today’s scientists must generate high‑quality, reliable data across a growing range of molecule types and workflows, often under increasing time pressure.

Posted by

Trained and working in Organic Synthesis I started using mass spectrometry to aid in my synthesis work. As my interest in mass spectrometry grew, I change roles several times within the pharmaceutical industry to be able to focus on its use and applications. I joined SCIEX in 2006, initially as a demo chemist, then with R&D, primarily focused on Analyst, and eventually SCIEX OS software development. More recently I switched to technical product management for Echo MS systems.

0 Comments

Submit a Comment

Pin It on Pinterest

Share This

Share this post with your network