Depending on the samples that you are running on the system, it is possible for the Echo MS electrode to become dirty or occluded over time. Below are two different cleaning strategies that will be helpful for you to maintain your system and keep your electrodes running well.
This protocol is easy to perform between running your sample plates.
This protocol is an offline protocol that can be used periodically for a more intensive wash of the electrode or to recover electrodes that are suffering from reduced flow rates.
Connect the OPI electrode to a small union (1/16 to 1/32) , IDEX Part # P-881 )
Some electrodes may not be able to be recovered but we have had good success with this method.
We recently hosted a webinar focused on streamlining forensic toxicology workflows, featuring expert speakers Maria Sarkisian from the San Francisco Office of the Chief Medical Examiner (SFOCME) and Dr. Dick Paul Kloos from the Netherlands Forensic Institute (NFI). The webinar explored innovative LC-MS/MS strategies that help forensic labs improve efficiency. In this blog, we share highlights from the Q&A session, where our speakers addressed the audience’s questions and shared actionable insights for forensic laboratory professionals.
We’re excited to launch our Ask the PFAS expert series, where we tackle some of the most pressing questions around PFAS testing, containment, and contamination control. In this first instalment, we sit down with Simon Roberts, a SCIEX application scientist, to share practical insights and expert advice.
Thanks to Starbucks, who launched the pumpkin spice latte in 2003 (yes, over 20 years ago), the spice mixture became a global phenomenon, loved and disliked at the same time.
Posted by
You must be logged in to post a comment.
Share this post with your network