Methods for electrode cleaning for Echo MS systems

Jun 3, 2025 | Blogs, Echo® MS+ system, Pharma | 0 comments

Read Time: 2 min

Depending on the samples that you are running on the system, it is possible for the Echo MS electrode to become dirty or occluded over time.  Below are two different cleaning strategies that will be helpful for you to maintain your system and keep your electrodes running well.

  1. Quick cleaning protocol

This protocol is easy to perform between running your sample plates.

  • From the Maintenance menu of the Echo MS direct control, select OPI Port Access then Lift to provide access to the top of the OPI port.
  • Disconnect electrode by loosening the finger tight red nut and pulling straight up
  • Place the OPI port end of the electrode into solvent. With the MS system still active, the draw created from the GS1 flow will draw solvent through the OPI into the MS source, turn Spray voltage to 0.
  • Wash with Milli-Q water for 1 min
  • Wash with 20% Acetonitrile with 0.1 M Formic Acid (0.4% FA) for 20 min
    • Drawing speed is about 450 µL/min based on the consumption of 20% Acetonitrile with 0.1 M FA
  • Wash with Milli-Q water for 1 min
  • Reattach the electrode to the system and tighten the red nut finger tight.
  • Restart the mobile phase flow and check the Maximum Supported Flow (MSF)

 

  1. Intensive cleaning protocol

This protocol is an offline protocol that can be used periodically for a more intensive wash of the electrode or to recover electrodes that are suffering from reduced flow rates.

Connect the OPI electrode to a small union (1/16 to 1/32) , IDEX Part # P-881 )

  • Place the electrode end of the capillary in a waste vessel capable of holding 200 mL or more
  • Flow at least 200 ml of 50% methanol through the electrode at a flow rate of 5 – 8 mL/min using an HPLC pump
  • Test on Echo MS system for droplet linearity, reproducibility, and peak shape
  • If storing the recovered electrode flush with 100% methanol before storing.

Some electrodes may not be able to be recovered but we have had good success with this method.

Why traditional MS/MS falls short for ADCs and how electron-activated dissociation (EAD) changes the equation

Antibody-drug conjugates (ADCs) are among the most promising—and analytically demanding—modalities in biopharma today. By combining large, heterogeneous antibodies with chemically labile payloads and linkers, ADCs introduce a level of structural complexity that challenges conventional analytical workflows. It also poses the challenge that the very features you need to measure are often the easiest to lose during analysis.

Six benefits of remote support every laboratory should know

Imagine having a tech expert at your fingertips to solve computer issues or a fitness trainer guiding you through workouts from the comfort of your home. In today’s fast-paced world, the ability to provide and receive service and support remotely is no longer a luxury but a necessity. Whether it’s troubleshooting a software issue, repairing a device, offering customer assistance, or enjoying the convenience of telehealth as a private individual, remote capabilities have revolutionized how businesses operate and how individuals get help

The costly consequences of unplanned downtime

Unplanned downtime is a formidable adversary that businesses across various industries strive to minimize. Defined as the unexpected interruption of regular operations, unplanned downtime can wreak havoc on productivity, profitability and customer satisfaction. In this article, we delve into the causes of unplanned downtime, its far-reaching consequences and strategies to mitigate its impact.

Posted by

Trained and working in Organic Synthesis I started using mass spectrometry to aid in my synthesis work. As my interest in mass spectrometry grew, I change roles several times within the pharmaceutical industry to be able to focus on its use and applications. I joined SCIEX in 2006, initially as a demo chemist, then with R&D, primarily focused on Analyst, and eventually SCIEX OS software development. More recently I switched to technical product management for Echo MS systems.

0 Comments

Submit a Comment

Pin It on Pinterest

Share This

Share this post with your network